ISOBLOCK V-4C (1000V 5V) 数据手册
IsoBlock V
Galvanicaly Isolated Differential Voltage Sensor
OVERVIEW
The IsoBlock V module has been designed to provide high-quality isolated differential voltage measurements for aplications requiring scaling of high voltages, as well as superior isolation.
Each IsoBlock V unit hosts four separate isolated channels, each of which can be connected to
separate measurement sources while providing a range of functional coverage up to 1500V. The
input of each specific IsoBlock channel has its own isolated reference, and can be configured to suit
user needs. All processed signals output from the IsoBlock unit are referenced in respect to the
ground channel of the user’s data acquisition system.
Verivolt designs its IsoBlock V modules with consideration for user great flexibility, high channeldensity and low power consumption.
SPECIFICATION
IsoBlock V
50V
100V
200V
Bandwidth (-3dB point)
300V
120kHz
Integrated sensor noise
(Referenced to input)
(1MHz custom)
< 2 mV
< 3 mV
< 6 mV
< 10 mV
< 14mV
5
10
50
75
100
150
100V
200V
400 V
600 V
1000 V
1500 V
2000 V
3000 V
4
Performance
Accuracy (percentage of reading)
Max total phase shift at 60Hz
Max Input delay (120kHz versions)
±0.2%
(±0.1% custom)
< 0.05°
< 2.8 μs
Input-Output non-linearity
< ±0.1%
Integrated channel noise
(Referenced to output)
< 1.2 mV
Output voltage
Gain temperature drift
Isolation voltage from primary side to
secondary side
> ±1500 V
Withstanding common mode surge
voltage
±5000 V
Withstanding differential mode surge
voltage
±2500 V
Mounting Type
DIN Rail
Connectivity (Connector for power in
and signal out to/from the sensor)
Common mode rejection
at 60Hz
Power Supply Voltage
Output type
Mechanical
Weight
1500V
< 1 mV
Eletrical
Channels
1000V
< 1 mV
Channels per module
Outer Dimensions
750V
< 1 mV
Gain
(Using 10V standard output voltage)
Differential input dynamic range
500V
Spring Cage connector
3.9” x 3.5” x 2.5”
1 and 4 channels
198 g (7.0 oz)
Output Offset Voltage
Differential Input
impedance
Insulation impedance
Output impedance
HARDWARE
DESCRIPTION
The IsoBlock V module is designed to isolate and scale down
high voltages found in industrial enviroments. The end result is
a signal ready to connect to any data acquisition system, while
galvanically isolating the source from it.
Each channel of the IsoBlock module has a galvanic isolation
from the input to the output that can eliminate large common
mode voltages. In addition to that, each channel also has a protection stage at the input that guards it from surges.
Following the input surge protection stage, there is an amplification stage that brings the input signal to a ±10V range. This signal
is modulated into a magnetic field, and then transferred across a
galvanic barrier. A demodulating stage recovers the original signal, followed by an anti-aliasing filter and a conditioning stage to
output a ±10V differential pair. The figure below shows a block
diagram of the process decribed above.
±10 V
±50 ppm/°C
112 dB
8V to 28 V
Differential pair
< ±500 µV
> 10 MΩ
IsoBlock V single channel block diagram.
> 10 GΩ || 2pF
100Ω
Environmental
Operating temperature
– 25 to 70 °C
Storage temperature
– 40 to 80 °C
- The isolation barrier of every device is tested with a 5 second partial discharge of 1800V for 5 seconds, with a detection threshold
of 150pC.
- Withstanding common mode surge voltage is 2 seconds half sinewave.
- Withstanding differential mode surge voltage is 4 seconds half sinewave.
650 Washington St, San Francisco, CA 94111
(708) 374-8658
1
MERCHANICAL
DIMENSIONS
HARDWARE
CONFIGURATION
A. Connect external power source
to power the unit. For proper functioning the power supply should provide a
voltage between 8V and 28V with at
least 0.25A of continuous current and
0.5A surge during module start-up.
A
Standards and Certifications
● CE
B. Securely connect wire in the 20-6
AWG range between the source of
measurement and an available IsoBlock’s input screw terminal.
C. Securely connect one end of a
twisted pair to the output terminals,
and the other end to the inputs of your
data acquisition unit
C
B
DANGER
THIS SENSOR IS NOT A SAFETY DEVICE AND IS NOT INTENDED TO BE USED AS A SAFETY DEVICE. This sensor is designed only to detect and read certain data in an electronic manner and perform no use apart from that, specifically no safetyrelated use. This sensor product does not include self-checking redundant circuitry, and the
failure of this sensor product could cause either an energized or de-energized output condition, which could result in death, serious bodily injury, or property damage.
2
ISOBLOCK V-4C (1000V 5V) 价格&库存
很抱歉,暂时无法提供与“ISOBLOCK V-4C (1000V 5V)”相匹配的价格&库存,您可以联系我们找货