TR220MB240-1538E12 VI 数据手册
TR220M SERIES
220 WATT MEDICAL
SWITCH ADAPTER
Features
Universal Input Range 80~264Vac
High Efficiency up to 94%
Class I (TR220MA), Class II (TR220MB)
No Load Input Power Consumption < 150mW
Approval IEC/EN/UL 60601-1 2 MOPP
Approval IEC/EN 60601-1-11 (TR220MB)
Approval EN55011 and CISPR/FCC Class B
Approval IP22
Operating Altitude 5000m
Continuous Short Circuit Protection
Over Voltage Protection
Meets CoC Tier 2 and DOE Level VI
MODEL
OUTPUT
OUTPUT
NUMBER
VOLTAGE
CURRENT
TR220MA120
TR220MA240
TR220MA360
TR220MA480
TR220MA560
TR220MB120
TR220MB240
TR220MB360
TR220MB480
TR220MB560
12 V
24 V
36 V
48 V
56 V
12 V
24 V
36 V
48 V
56 V
16.67 A
9.17 A
6.11 A
4.58 A
3.93 A
16.67A
9.17 A
6.11 A
4.58 A
3.93 A
RIPPLE &
VOLTAGE
LINE
LOAD
%EFF.
NOISE
ACCURACY
REGULATION
REGULATION
(Typ.)
NOTE1
NOTE2
NOTE3
NOTE4
NOTE5
120mV
200mV
200mV
200mV
200mV
120mV
200mV
200mV
200mV
200mV
±2%
±2%
±2%
±2%
±2%
±2%
±2%
±2%
±2%
±2%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±4%
±3%
±2%
±2%
±2%
±4%
±3%
±2%
±2%
±2%
91%
93%
93%
94%
94%
91%
93%
93%
94%
94%
Note:
1. Add a 0.1uF ceramic capacitor and a 10uF E.L. capacitor to output for ripple & noise measuring @20MHz BW.
2. Voltage accuracy is set at 60% full load.
3. Line regulation is measured from 100Vac to 240Vac with full load.
4. Load regulation measured from 60% to 100% full load and from 60% to 20% full load (60%±40% full load).
5. Typical efficiency at 230Vac and 75% full load at 25℃.
PART NUMBER
Series
TR220M
X
Output Voltage
DC Plug Type
Cable Type
Cable Length
XXX
-XX
E
XX
120:12V
220W
Medical
Adapter
A:Class I
B:Class II
240:24V
360:36V
471:950mm with Ferrite Core
See Page 6
E:UL2464 with OVP
12:1220mm with Ferrite Core
480:48V
13:1800mm with DC Jack
560:56V
See page 6 restrictions
Part Number Example:
TR220MA240-1538E12, 220W, Class I, 24Vdc Output, Din Power Plug Molded Type, Cable Length 1220mm
TR220MB360-11E13, 220W, Class II, 36Vdc Output, DC Jack Type, Cable Length 1800mm
www.cincon.com
1
REVISION: V10
TR220M
Series
TECHNICAL SPECIFICATIONS
(All specifications are typical at nominal input, full load at 25℃ unless otherwise noted.)
ABSOLUTE MAXIMUM RATINGS
PARAMETER
NOTES and CONDITIONS
Input Voltage
Operating Temperature
See Derating Curve
Storage Temperature
Input/Output Isolation Voltage
1 minute
Operating Altitude
Device
Min.
Typ.
Max.
Units
Vac
All
80
264
All
-30
70
All
-40
85
All
4000
Vac
All
5000
m
Max.
Units
240
Vac
63
Hz
INPUT CHARACTERISTICS
PARAMETER
Operating Voltage Range
Input Frequency Range
Maximum Input Current
Min.
All
100
All
47
Typ.
All
2.5
A
Leakage Current (Earth)
All
300
uA
Leakage Current (Touch)
All
Under Voltage Protection
All
50
All
0.92
Power Factor
100% Load, Vin=100Vac
Device
230Vac/50Hz @ Full load
100
uA
60
65
V
Units
OUTPUT CHARACTERISTICS
PARAMETER
Output Voltage Set Point
Operating Output Current Range
Holdup Time
Vin=115Vac and 230Vac, Io=60% Full load
Tc=25℃
Vin=115Vac and 230Vac,Tc=25℃
Vin=115Vac
Device
Min.
Typ.
Max.
TR220MA/B120
11.76
12
12.24
TR220MA/B240
23.52
24
24.48
TR220MA/B360
35.28
36
36.72
TR220MA/B480
47.04
48
48.96
TR220MA/B560
54.88
56
57.12
TR220MA/B120
TR220MA/B240
TR220MA/B360
TR220MA/B480
TR220MA/B560
All
Vdc
A
ms
Output Voltage Regulation
Load Regulation
Line Regulation
Output Ripple and Noise
www.cincon.com
60%±40% Full load change
Vin=High line to low line
TR220MA/B120
±4
TR220MA/B240
±3
TR220MA/B360
±2
TR220MA/B480
±2
TR220MA/B560
±2
All
Latch Off
10uF aluminum electrolytic capacitor
to output
2. Oscilloscope is 20MHz band width
3. Ambient temperature=25℃
2
±1
TR220MA/B120
13.5
TR220MA/B240
30
TR220MA/B360
42
TR220MA/B480
56
TR220MA/B560
59.2
Vdc
TR220MA/B120
120
TR220MA/B240
200
TR220MA/B360
200
TR220MA/B480
200
TR220MA/B560
200
mV
REVISION: V10
TR220M
PARAMETER
Load Capacitance
Efficiency
1. Vin=115Vac and 230Vac
2. Output is max. load
1. Vin=230Vac
2. Output is 75% full load
Device
Min.
Typ.
Series
Max.
TR220MA/B120
16500
TR220MA/B240
9270
TR220MA/B360
6160
TR220MA/B480
4540
TR220MA/B560
4070
TR220MA/B120
91%
TR220MA/B240
93%
TR220MA/B360
93%
TR220MA/B480
94%
TR220MA/B560
94%
Units
uF
ISOLATION CHARACTERISTICS
PARAMETER
Input to Output
1 minute (without dielectric breakdown)
Input to Earth (Ground)
Output to Earth (Ground)
Isolation Resistance
Device
Min.
Typ.
Max.
Units
All
4400
Vac
All
1800
Vac
All
1800
Input to output
All
100
Device
Min.
Vac
MΩ
FEATURE CHARACTERISTICS
PARAMETER
Switching Frequency
All
Typ.
Max.
100
Units
kHz
GENERAL SPECIFICATIONS
PARAMETER
Device
Min.
MTBF
Io=100%; Ta=25℃ per MIL-HDBK217F
All
585
Life Time
@ 75% Load, 40℃
All
26
Humidity
Non-condensing
MIL-STD-810F Table 516.5,TABLE
516.5-I 10ms, each axis 3 times(±X、
±Y、±Z axis)
MIL-STD-810F Table 514.5C-VIII,
15~2000Hz, X、Y、Z axis, 1
hour(each axis),. total 3 hours.
All
Shock
Vibration
Typ.
Max.
Units
93
k
hours
k
hours
% RH
All
75
g
All
4
g
Weight
All
Dimension
All
950
7.087x2.953x1.752 Inches
(180x75x44.5mm)
grams
Ed 3.1
EMC Emission
Class I, Class II, IEC60601-1, EN60601-1, ANSI/AAMI ES60601-1
IEC/EN 60601-1-11 for TR220MB (Home Health Care)
EN55011:2016+A1:2017, EN61000-3-2:2014, EN6100-3-3:2013, FCC CFR47 Part 15
Conducted Disturbance
Radiated Disturbance
EN55011:2016+A1:2017, FCC CFR47 Part 15
Class B
Class B
Harmonic Current Emissions
EN 61000-3-2:2014
Class A
Voltage Fluctuations & Flicker
EN 61000-3-3:2013
Criterion A
EMC Immunity
EN60601-1-2:2015, IEC61000-4-2, 3, 4, 5, 6, 8, 11
Electrostatic Discharge (ESD)
Radio-Frequency, Continuous
Electrical Fast Transient (EFT)
IEC 61000-4-2:2008 Air Discharge: ±15kV Contact Discharge: ±8kV
Criterion A
IEC 61000-4-3:2020
Criterion A
IEC61000-4-4:2012, ±1kV, ±2kV
Criterion A
Surge
Conducted Disturbances, Induced
by RF Fields
Power Frequency Magnetic Field
IEC61000-4-5:2014+A1:2017, L-N: ±0.5kV, ±1kV, ±2kV L-E (Ground): ±0.5kV, ±1kV, ±2kV
Criterion A
IEC 61000-4-6:2013+COR1:2015
Criterion A
IEC 61000-4-8:2009
Criterion A
Voltage Dips
IEC 61000-4-11:2004+A1:2017, Dips:30% reduction, Dips: >95% Reduction, Criteria A
Criterion A
Voltage Interruptions
IEC 61000-4-11:2004+A1:2017, >95% Reduction
Criterion B
Safety
Application Note Link
www.cincon.com
Ed 4.0
TR220M Series App Notes
3
REVISION: V10
TR220M
Series
CHARACTERISTIC CURVE
Power Derating Curve
TR220M Derating Curve
TR220M Input Voltage Derating Curve
100%
100%
Rated Load (%)
120%
Rated Load (%)
120%
80%
60%
40%
20 LFM (Natural Convection)
20%
0%
80%
60%
40%
20%
0%
-30
-20
-10
40
50
60
70
80
100
120
Ambient Temperature (℃)
140
160
180
200
220
240
260
Input Voltage (Vac)
Performance Data
TR220M240 (Eff Vs Io)
95%
95%
90%
90%
Efficincy (%)
Efficincy (%)
TR220M120 (Eff Vs Io)
85%
80%
75%
115V
70%
230V
80%
75%
115V
70%
230V
65%
60%
10%
85%
65%
20%
30%
40%
50%
60%
70%
80%
60%
10%
90% 100%
20%
30%
Current Load (%)
50%
60%
70%
80%
TR220M480 (Eff Vs Io)
95%
90%
90%
Efficincy (%)
95%
85%
80%
75%
115V
70%
85%
80%
75%
115V
70%
230V
230V
65%
60%
10%
90% 100%
Current Load (%)
TR220M360 (Eff Vs Io)
Efficincy (%)
40%
65%
20%
30%
40%
50%
60%
70%
80%
60%
10%
90% 100%
Current Load (%)
20%
30%
40%
50%
60%
70%
80%
90% 100%
Current Load (%)
TR220M560 (Eff Vs Io)
95%
Efficincy (%)
90%
85%
80%
75%
115V
70%
230V
65%
60%
10%
20%
30%
40%
50%
60%
70%
80%
90% 100%
Current Load (%)
www.cincon.com
4
REVISION: V10
TR220M
Series
MECHANICAL SPECIFICATION
All Dimensions are in inches[mm]
Tolerance:Inches:X.XXX±0.02
Millimeters:X.XX±0.5
UNIT:inches[mm]
Din Power Plug
7.087[180.00]
37.402[950.00]±1.969[50.00]
LED
IEC320/C14
2.953[75.00]
P1 P3
P2 P4
Pin1,3:- Pin2,4:+ Shell:NC
4Pin Power Plug Assembly Type
16AWG/950mm
1.713[43.50]
IEC320/C8
DC Jack
7.087[180.00]
70.866[1800.00]±1.969[50.00]
LED
2.953[75.00]
IEC320/C14
DC Plug type: V+ ─●)─ VDC Plug :Straight(ψ5.5/ψ2.5) L12mm
16AWG/1800mm
1.713[43.50]
IEC320/C8
* For Output Voltage 12Vdc and 24V model, it must select Din Power Plug Molded Type or equivalent
* For Output Voltage 36Vdc to 56Vdc models, it’s able to select Din Power Plug Molded Type, DC Jack or equivalent.
www.cincon.com
5
REVISION: V10
TR220M
Series
Cable Type
Cable Length
Cable AWG
UL2464
1800mm with Ferrite Core
Cable Type
Cable Length
STANDARD OUTPUT DC PLUG
DC Plug Type
A
B
C
OD
(mm)
ID
(mm)
L
(mm)
11E13
Φ5.5
Φ2.1
12
12E13
Φ5.5
Φ2.5
12
23E13
Φ5.5
Φ2.1
9.5
26E13
Φ5.5
Φ2.5
9.5
01E13
Φ5.5
Φ2.1
12
02E13
Φ5.5
Φ2.5
12
21E13
Φ5.5
Φ2.5
9.5
24E13
Φ5.5
Φ2.1
9.5
Cable NumberXXXXX
16AWG for
36V,48V,56V
Pin Assignment
Din Plug Type
KYCON KPPX-4P equivalent
with Lock (Din Power Plug
Assembly Type)
Cable Number XXXXX
1446E12
1446E471
without Lock (Din Power Plug
Molded Type)
1538E12
1538E471
PIN No.
Polarity
P1
P2
P3
P4
Shell
No Connection
P1
P2
P3
P4
Shell
No Connection
Cable AWG
1200mm with Ferrite Core 16AWG for 24V
UL2464
16AWG for 12V
UL2464
16AWG for 12V
※Other DC Plug Type please refer to the link: https://www.cincon.com/productdownload/TR220M-cable-DC-plug.pdf
6
REVISION: V10