Datasheet SEN5x
Environmental Sensor Node for HVAC and Air Quality Applications
PM, NOx, VOC, RH & T sensor platform
Fast & easy integration
One driver for up to 8 data signals
Superior sensing accuracy and lifetime
Fully calibrated digital output
Overview
The SEN5x is a unique sensor module family combining the measurement of critical air quality parameters –
particulate matter, VOC, NOx, humidity, and temperature – in a single package. The modules are a result of
Sensirion’s extensive experience in environmental sensing and offers best possible performance of each
sensing parameter, a superior lifetime, as well as a compact form factor. The SEN5x sensors provide fully
calibrated outputs with a digital interface. The combination of all measurement parameters in a single device
allows manufacturers of air quality devices to reduce design and integration efforts, shorten development times
and simplify their supply chain. Proprietary algorithms enable straightforward integration into HVAC and air
quality applications allowing resource/time savings and focus on core competencies. The implemented
Sensirion Temperature Acceleration Routine (STAR) engine accelerates the device’s response to ambient
temperature change by a factor of 2-3 to provide better UX and more accurate measurements to end users.
www.sensirion.com
Version 2 – D1 – March 2022
1/30
Air
Flow
RH & T
I2C
VOC & NOx
3V3
Laser
Power
Supply
5V
VDD
Laser
Controller
SDA
MCU
SCL
Amplifier &
Filter
Photodiode
Fan
ESD
Protection
GND
Fan
Controller
Figure 1 Functional block diagram of the SEN5X. The RH & T and VOC sensors are only contained in the SEN54 and SEN55, the
NOx signal is only contained in the SEN55.
www.sensirion.com
2/30
Content
1 Product Family Overview and Ordering Information
4
2 Environmental Sensor Node Specifications
5
3 Electrical Specifications
11
4 Hardware Interface Specifications
13
5 Functional Overview
6 Operation and Communication through the
14
I2C
Interface
17
7 Technical Drawings
26
8 Revision History
28
9 Important Notices
29
www.sensirion.com
3/30
The SEN5x product family consists of a lineup of sensors with identical dimensions, identical hardware & firmware
interface, and varying number of sensor outputs. Table 1 gives an overview of all available product variants with their
respective sensor outputs.
Specifications and commands for sensor inputs/outputs only apply for the given product if the sensor output is part of
this product.
Products
SEN50-SDN-T
Sensor outputs
Particulate Matter
Description
Article Number
3.000.667
Particulate matter sensor, tray packaging,
45 pcs. per tray
SEN54-SDN-T
Particulate Matter
Relative Humidity
Temperature
VOC Index
Environmental sensor node, tray packaging,
45 pcs. per tray
3.000.535
SEN55-SDN-T
Particulate Matter
Relative Humidity
Temperature
VOC Index
NOx Index
45 pcs. per tray
3.000.593
Table 1: Sensor module specifications.
www.sensirion.com
4/30
Default conditions of standard Measurement-Mode (6.1.1), 25 °C and 5 V supply voltage apply to values in the table
below, unless otherwise stated.
2.1 Sensor Module Specifications
Parameter
Sampling interval
Conditions
Lifetime1
Acoustic emission level
Long term acoustic emission level drift
Weight
24 h/day operation
0.2 m
0.2 m
Value
1 0.03
> 10
24
+0.5
36.4 10%
Units
s
years
dB(A)
dB(A) / year
g
Table 2: Sensor module specifications.
1
Lifetime is based on mean-time-to-failure (MTTF) calculation. Lifetime might vary depending on different operating conditions.
www.sensirion.com
5/30
2.2 Particulate Matter Specifications
Parameter
Mass concentration specified range
Mass concentration size range
Conditions
PM1.0
PM2.5
PM4
PM10
0 to 100 μg/m3
Value
0 to 1’000
0.3 to 1.0
0.3 to 2.5
0.3 to 4.0
0.3 to 10.0
Units
μg/m3
μm
μm
μm
μm
100 to 1000 μg/m3
5 μg/m3 AND 5 % m.v.
% m.v.
10
0 to 100 μg/m3
25
μg/m3
100 to 1000 μg/m3
25
% m.v.
Maximum long-term mass concentration precision
limit drift
0 to 100 μg/m3
1.25
μg/m3 / year
Typical start-up time6
number
concentration
Mass concentration precision2,3 for PM1 and PM2.54
Mass concentration
precision2,3
for PM4,
PM105
Sensor output characteristics
Additional T-dependent mass precision limit drift3
Laser wavelength
(DIN EN 60825-1 Class 1)
100 to 1000
μg/m3
200 – 3000 #/cm3
100 – 200 #/cm3
50 – 100 #/cm3
PM2.5 mass concentration
temperature
difference to 25°C
typ.
typ.
% m.v. / year
1.25
8
s
16
s
30
s
Calibrated to TSI DustTrak™
DRX 8533 Ambient Mode
% m.v. / °C
0.5
660
nm
Table 3: Particulate matter sensor specifications. Default conditions of 252 °C, 5010% relative humidity and 5 V supply voltage
apply unless otherwise stated. ‘max.’ means ‘maximum’, ‘typ.’ means ‘typical’, ‘% m.v.’ means ‘% of measured value’.
2
Also referred to as “between-parts variation” or “device-to-device variation”.
For further details, please refer to the document “Sensirion Particulate Matter Sensor Specification Statement”.
4 Verification Aerosol for PM2.5 is a 3% atomized KCl solution. Deviation to reference instrument is verified in end-tests for every sensor after calibration.
5 PM4 and PM10 output values are calculated based on distribution profile of all measured particles.
6 Time after starting Measurement-Mode, until a stable measurement is obtained.
3
www.sensirion.com
6/30
2.1 Temperature and Humidity Specifications
Parameter
Compensated outputs8
Conditions
Value 7
Temperature and
Relative Humidity
Typical accuracy temperature
@ 15-30 °C, 50 %RH
Max. accuracy temperature
Repeatability temperature
Response time temperature9
@ 15-30 °C, 50 %RH
@ 25 °C, 50 %RH
0.45
±0.7
0.1
< 60
Typical accuracy relative humidity
@ 25 °C, 30-70 %RH
Max. accuracy relative humidity
Repeatability relative humidity
@ 25 °C, 30-70 %RH
@ 25 °C, 50 %RH
Response time relative humidity 10
@ 25 °C, 50 %RH, 63%
4.5
±6
1
< 20
Units
°C
%RH
°C
°C
°C
s
%RH
%RH
%RH
s
Table 4: Temperature and humidity specifications.
7
For the definition of the typical and max. accuracy tolerance, please refer to the document “Sensirion Humidity Sensor Specification Statement”.
Self-heating of the module is compensated according to the application note “Temperature Acceleration and Compensation Instructions for SEN5x”.
9 For a step from 15°C to 25°C
10 For a step from 75%RH to 25%RH
8
www.sensirion.com
7/30
2.2 Gas Specifications
Parameter
Output signals
Device-to-device
variation
Values
Min.
1
1
Comments
VOC Index
NOx Index
VOC Index12
Typ.11
很抱歉,暂时无法提供与“SEK-SEN5X”相匹配的价格&库存,您可以联系我们找货
免费人工找货- 国内价格 香港价格
- 1+617.944881+79.83366