2007-05-23
Ambient Light and Proximity Sensor with Integrated 940nm IR Emitter
Umgebungslicht- und Proximity Sensor mit integriertem 940nm IR Emitter
Version 1.0
SFH 7779
Features:
• Proximity sensor (PS)
- Detection range up to 160 mm
- 940 nm IR emitter integrated in package
- Programmable pulse current up to 200 mA
- No separator needed if placed 100mA
ILED
25
Accuracy of LED current source
(ALS_PS_CONTROL: LED Current = 00b)
ILED
22.5
Mean current consumption in PS mode
(current consumption of the pulsed LED is not
included; MODE_CONTROL(0x41h) = 0x03h;
all other registers are default; VDD = 2.5V)
300
µs
200
mA
25
27.5
mA
IDD
60
150
µA
Mean current consumption in PS mode during
the 200µs LED pulse (t LED ON)
included)
IDD
6.5
8.5
mA
Typical detection distance 1)
KODAK grey card 100x130mm², R=90%
(ILED=200mA, VDD=2.5V, Ev = 0lx;
high threshold = 10 counts)
hon
10
2014-07-25
7
200
cm
Version 1.0
SFH 7779
Typical none detection distance 1)
low threshold = 7 counts)
hoff
12
cm
PS sensor output with human skin reflector
(ILED=200mA, VDD=2.5V, h=0mm: skin directly
on top of the sensor)
PSout
850
counts
Temperature coefficient of the PS signal
(ILED=200mA, VDD=2.5V, R=90%,
AReflector =10x13cm², h=4cm)
TKPS
0.15
%/K
1)
Example of PS Hysteresis
The switching distance h is specified from top sensor surface to the reflector.
INT voltage
Reflector:
(Kodak grey card R=90%)
HIGH
h
h,HYS
SFH 7779
LOW
h_on
2014-07-25
h_off
Reflector
distance h
8
Version 1.0
SFH 7779
Dimensions of proposed optical aperture
(optical aperture: IR transmitting opening in cover glass)
IR Intransparent and Low
Reflective Cover
(deep black)
Cover Glass
Ø Aperture
35°
Ø Aperture
35°
35°
35°
Dd
SFH 7779
PS & ALS
Detector
0.45
pcb
0.8
0.675
,5WUDQVSDUHQWRSHQLQJ
6)+
0.9
Emitter
36 $/6
'HWHFWRU
,5,QWUDQVSDUHQWDQG/RZ
5HIOHFWLYH&RYHU
GHHSEODFN
(PLWWHU
3LQ
UHFRPPHQGHGDSHUWXUHG PP
Dimensions in mm
2014-07-25
9
Version 1.0
SFH 7779
Characteristics (Ta = 25°C)
Parameter
Symbol
Value
min.
Unit
typ.
max.
Ambient Light Sensors: ALS VIS and ALS IR diode
Wavelength of max. sensitivity for ALS VIS
Spectral range of sensitivity (10% of Smax) of
ALS VIS
Wavelength of max. sensitivity of ALS IR
ALS IR
Smax
S10%
Smax
S10%
Illuminance measurement range is
programmable
(the maximum ALS sensitivity can be reached
with the ALS high sensitivity mode --> 400ms
ALS integration time)
(MODE_CONTROL (0x41h) = 0Ah or 0Bh)
560
450
nm
950
880
nm
nm
830
1050
nm
0.0022
73000
lx
ALS VIS sensor output
(1000lx; white LED; VDD = 2.5V)
((MODE_CONTROL (0x41h) = 08h)
(ALS_PS_CONTROL (0x42h): Gain = X1)
ALSVIS_out
750
900
1080
counts
ALS IR sensor output
(324µW/cm²; IRED 850 nm; VDD = 2.5V)
ALSIR_out
460
550
660
counts
ALS VIS sensor output at darkness
ALSVIS_out
counts
ALS IR sensor output at darkness
ALSIR_out
counts
2014-07-25
10
Version 1.0
Parameter
SFH 7779
Symbol
Value
min.
Resolution of the digital output signal based on
gain settings for ALS VIS:
tint ALS = 100ms
Gain X1
Gain X2
Gain X 64
Gain X 128
Unit
typ.
ALSVIS_out
lx/count
1.1
0.55
0.018
0.009
High sensitive mode:
tint ALS = 400ms
Gain X 128
(VDD = 2.5V; white LED)
- Gain settings at ALS_PS_CONTROL (0x42h)
0.002
Typical temperature coefficient for ALS
measurement
TCEv
0.2
(other registers are in default)
IDD
90
Typical error by Flicker noise
(caused by bulbs (f=50 or 60Hz) or fluorescent
lamps)
2014-07-25
max.
11
%/K
150
µA
3
Version 1.0
SFH 7779
Diagrams for ALS sensor
Relative Spectral Sensitivity of ALS VIS
Srel_VIS = f(λ)
100
Srel %
Relative Spectral Sensitivity of ALS IR
Srel_IR = f(λ); 100% = maximum sensitivity of ALS VIS
diode
OHF05597
100
Srel %
80
80
70
70
60
60
50
50
40
40
30
30
10
400 500 600 700 800 900 nm 1100
λ
2014-07-25
OHF05606
λ
12
Version 1.0
SFH 7779
ALS VIS sensitivity ranges
ALS VIS output f(Ev); white LED; f(sensitivity settings);
Tint: integration time (register 0x41h); X: gain settings (register 0x42h)
OHF05598
100000
10000
X = 64
ALS Vis out / counts
Tint = 100 ms
X = 128
1000
Tint = 100 ms
X=2
X = 128
Tint = 100 ms
Tint = 400 ms
100
X=1
Tint = 100 ms
10
1
0.001
0.1
10
1000
lx
100000
EV
Direction Characteristic of ALS Vis diode
Srel f(f))
40˚
30˚
20˚
10˚
ϕ
0˚
OHF05605
1.0
50˚
0.8
60˚
0.6
70˚
0.4
80˚
0.2
90˚
100˚
2014-07-25
1.0
0.8
0.6
0.4
0˚
13
20˚
40˚
60˚
80˚
100˚
120˚
Version 1.0
SFH 7779
Diagrams for PS sensor
PS sensitivity f(R = reflectivity): 4%; 18%; 90%;
VDD=2.5V; ILED=200mA; Trep=100ms; AReflector = 10 x 13 cm²
PS sensor output / counts
1000
100
10
1
100
150
distance h / mm
2014-07-25
14
200
250
Version 1.0
SFH 7779
PS sensitivity f(ILED): 200mA; 100mA; 50mA;
VDD=2.5V; Trep=100ms; R=90%; AReflector = 10 x 13 cm²
1000
100
10
1
100
150
200
250
distance h / mm
Typical crosstalk free range
VDD=2.5V; ILED=200mA; Trep=100ms; glass without ink and one single IR transmissive opening ;
OHF05599
1.2
mm
1.0
Distance cover bottom surface
to top of the sensor
> 1 count
0.8
1 count
0.6
0 counts
0.4
0.2
0.4
0.6
15
0.8
Thickness of coverglass
1
mm
1.2
Version 1.0
SFH 7779
VDD=2.5V; ILED=200mA; Trep=100ms; glass without ink and a two separate IR transmissive openings for
emitter and detector
OHF05609
1.2
> 1 count
mm
1.0
1 count
0.8
0 counts
0.6
0.4
0.2
0.4
0.6
0.8
1
mm
1.2
Note for crosstalk free range: The displayed crosstalk free ranges for a single or two hole IR transmissive opening are
measured with a clear cover.Depending on the used ink the crosstalk level can differ and needs to be measured.OSRAM
OS provides costumer related application support and measurements - please contact your OSRAM OS marketing or
sales partner, if support is required.
As the measurement results show the typical performance of the sensor OSRAM OS recommends to design inn a
additional safety guard in the distance of the cover bottom surface to the top of the sensor of 200µm.
e.g.: for a 0.5mm thick cover window with two holes and an acceptable crosstalk level of 0 counts the distance of the
cover window bottom surface to the top of the sensor should not exceed 0.6mm (typ. it would be 0.8mm)!
2014-07-25
16
Version 1.0
SFH 7779
Radiation Characteristics of the IR Emitter
Irel f(f)
40˚
30˚
20˚
10˚
ϕ
0˚
OHF05605
1.0
50˚
0.8
60˚
0.6
70˚
0.4
80˚
0.2
90˚
100˚
2014-07-25
1.0
0.8
0.6
0.4
0˚
17
20˚
40˚
60˚
80˚
100˚
120˚
Version 1.0
SFH 7779
Diagrams for IDD current consumption
Current consumption IDD in standby mode
IDD = f(VDD); Register 0x41= 0x00
IDD
Current consumption IDD in PS mode
IDD = f(VDD); Register 0x41= 0x03
OHF05602
4.5
µA
OHF05603
120
µA
IDD
100
3.5
80
3.0
2.5
60
2.0
40
1.5
0.5
2.2
2.5
2.8
3.1
3.4 V 3.7
VDD
OHF05604
IDD
120
100
80
60
40
2.5
2.8
3.1
3.4 V 3.7
VDD
2014-07-25
2.8
3.1
3.4 V 3.7
VDD
Current consumption IDD in ALS mode
IDD = f(VDD); Register 0x41= 0x08
160
µA
2.5
18
Version 1.0
SFH 7779
Register
SYSTEM_CONTROL register (0x40h)
The SYSTEM_CONTROL register is used to control the software (SW) reset and the interrupt function
(INT). Manufacturer ID and Part ID can be read.
R/W-Register 0x40
Bit
7
6
SW reset
INT reset
default 0 Initial reset is not
started
0 INT pin status is not
initialized
5
4
3
Manufacturer
ID
(Read only)
001
2
1
Part ID
(Read only)
001
0 Initial reset is not started 0 INT pin status is not initialized
1 Initial reset started
1 INT pin become inactive
(high impedance )
MODE_CONTROL register (0x41h)
CONTROL of PS and ALS operating modes and time settings.
Repetition time is the time between two separate measurements. Integration time is the duration for one
measurement. ALS high sensitivity modes are 1010 and 1011 with an increased integration time of
400ms. In PS operating mode: „normal mode“ only one PS measurement is performed during one PS
repetition time. In PS operating mode „twice mode“ two independent PS measurement are performed
within one PS repetition time. Both measurements are independent and can trigger the interrupt. This
feature can be used to decrease the interrupt update time if the persistence function (register 0x43h) is
used.
2014-07-25
19
Version 1.0
SFH 7779
R/W-Register 0x41
Bit
7
6
5
4
3
2
1
Reserved PS operating mode
default
Repetition /
Integration
time
Repetition time
ALS
PS
0 normal mode
0000
standby
standby
0 normal mode
0000
standby
standby
1 twice mode
0001
standby
10ms
0010
standby
40ms
0011
standby
100ms
0100
standby
400ms
0101
100ms / 100ms
standby
0110
100ms / 100ms
100ms
0111
100ms / 100ms
400ms
1000
400ms / 100ms
standby
1001
400ms / 100ms
100ms
1010
400ms / 400ms standby
1011
400ms / 400ms 400ms
1100
50ms / 50ms
50ms
Rest forbidden
ALS_PS_CONTROL register (0x42h)
ALS and PS Control of set the PS output mode, the ALS gain and the LED current. In the „Infrared DC
level output“ PS mode (bit = 1) the sensor measures the infrared DC ambient level. The proximity
value of the reflected signal is not available in this mode.
R/W-Register 0x42
Bit
7
6
Reserved
(read only)
PS output
default
2014-07-25
write 0
0 proximity output
5 4 3 2
1
ALS Gain for ALS VIS and ALS
IR
0000
ALS VIS
ALS PS
X1
X1
LED current
11 200mA
Version 1.0
SFH 7779
R/W-Register 0x42
Bit
7
6
5 4 3 2
Reserved
(read only)
PS output
1
IR
LED current
0 proximity output
0000
X1
X1
00 25 mA
1 Infrared DC level
output
0100
X2
X1
01 50 mA
0101
X2
X2
10 100 mA
1010
X64
X64
11 200 mA
1110
X128
X64
1111
X128
X128
rest
forbidden
Persistence Register (0x43h)
Settings for the interrupt activation and of the persistence interrupt function. Persistence function is only
valid for the PS measurements.
R/W-Register 0x43
Bit
7
6
5
4
3
2
1
Reserved (read only) Persistence
default 0000
2014-07-25
0001 Interrupt status is updated after each measurement
21
Version 1.0
SFH 7779
R/W-Register 0x43
Bit
7
6
5
4
3
2
1
0000 Interrupt becomes active after each measurement
(The mode indicates that a PS or ALS measurement has been finished and can
be read via the register. It is independent of the ALS & PS measurement value
and threshold settings)
(The interrupt status is updated independently after each measurement. Active
or Inactive status of the interrupt is depending on the values of the last
measurement in combination with the interrupt settings :“interrupt mode“
(register 0x4Ah) and „thresholds“ register 0c4Ch and following.)
0010 Interrupt status is updated if two consecutive threshold
judgement are the same
(The interrupt status only changes if the interrupt judgement of 2 consecutive
measurement results are the same and different to the current interrupt status.)
0011 ... 1111 Interrupt status is updated if threshold judgement are
the same over consecutive set times (3 .. 15)
(This is the same procedure like in the 0010 persistence mode, but instead of
2 consecutive threshold judgments more are needed (3 to 15 depending on the
setting) to change the interrupt status.)
e.g.:
1010: 10 measurement results in a row need to fulfill the interrupt judgement to
update the interrupt status
PS_DATA_LSBs register (0x44h)
LSB of the PS output.
R-Register 0x44
Bit
7
27
default 0
6
5
4
3
2
1
25
24
23
22
21
PS_DATA_MSBs register (0x45h)
MSB of the PS output.
R-Register 0x45
Bit
7
6
5
4
not used
not used
not used
not used
211
210
29
28
default 0
2014-07-25
22
3
2
1
Version 1.0
SFH 7779
ALS_VIS_DATA_LSBs register (0x46h)
LSB of the ALS VIS output.
R-Register 0x46
Bit
7
27
default 0
6
5
4
3
2
1
25
24
23
22
21
ALS_VIS_DATA_MSBs register (0x47h)
MSB of the ALS VIS output.
R-Register 0x47
Bit
7
215
default 0
6
5
4
3
2
1
213
212
211
210
29
28
ALS_IR_DATA_LSBs register (0x48h)
LSB of the ALS IR output.
R-Register 0x48
Bit
7
27
default 0
6
5
4
3
2
1
25
24
23
22
21
ALS_IR_DATA_MSBs register (0x49h)
MSB of the ALS IR output.
R-Register 0x49
Bit
7
215
default 0
2014-07-25
6
5
4
3
2
1
213
212
211
210
29
28
Version 1.0
SFH 7779
Interrupt function setting register (0x4Ah)
Setting of the interrupt functions.
R/W-Register 0x4A
Bit
7
6
PS INT
status
ALS INT
status
5
4
PS INT mode
3
2
1
INT assert
INT latch
INT trigger
(read only) (read only)
default 0 inactive
0 inactive
00 PS_TH is only
active
0 INT „L“ is stable
0 INT is latched
00 inactive
0 inactive
0 inactive
00 PS_TH (PS high
threshold 0x4Bh &
0x4Ch) is only active
0 INT „L“ is stable if
newer measurement
results is also
interrupt active
0 INT is latched
until INT
register is read
or initialize
00 INT pin is
inactive
1 active
1active
01 PS_TH & PS_TL
(PS high & low
threshold) are active
as hysteresis
1 INT “L“ is de-assert
and re-assert if
newer measurement
results is also
interrupt active
1INT is updated
after each
measurement
01 triggered
by PS only
10 PS_TH & PS_TL
(PS high & low
as outside detection
10 triggered
by ALS only
11 forbidden
11 triggered
by PS or ALS
PS INT and ALS INT status (bit ): Directly after reading the register the interrupt status for PS and
ALS and the INT Pin of the sensor is automatically set back to inactive status independent on the measurement results.
PS INT mode (bit ): The INT modes are only valid for the PS interrupt function. For description
please see extra chapter „PS INT Modes“ (at the end of the register chapter).
INT assert (bit ): Is used to adjust the sensor behaviour to the used micro controller trigger settings.
In case a repeated trigger in low state is needed the INT assert can be set to 1.
INT trigger (bit ): defines the source / sources for the interrupt.
INT latched (bit ): In latched mode the interrupt status stays active after the first activation. It is only
released by reading the status are performing an interrupt reset.
PS_TH_LSBs register (0x4Bh)
LSB for the PS threshold „HIGH“.
R/W-Register 0x4B
Bit
7
7
6
2
2
default 1
1
2014-07-25
6
5
2
1
5
4
4
3
2
2
1
1
24
3
2
2
1
1
2
2
1
1
1
Version 1.0
SFH 7779
PS_TH_MSBs register (0x4Ch)
MSB for the PS threshold „HIGH“.
R/W-Register 0x4C
Bit
7
default 0
6
2
1
210
29
28
1
1
1
1
PS_TL_LSBs register (0x4Dh)
LSB for the PS threshold „LOW“.
R/W-Register 0x4D
Bit
7
27
default 0
6
5
4
3
2
1
25
24
23
22
21
PS_TL_MSBs register (0x4Eh)
MSB for the PS threshold „LOW“.
R/W-Register 0x4E
Bit
7
default 0
6
2
1
210
29
28
ALS_VIS_TH_LSBs register (0x4Fh)
LSB for the ALS VIS threshold „HIGH“.
R/W-Register 0x4F
Bit
7
27
default 1
6
5
4
3
2
1
25
24
23
22
21
1
1
1
1
1
1
ALS_VIS_TH_MSBs register (0x50h)
MSB for the ALS VIS threshold „HIGH“.
R/W-Register 0x50
Bit
7
215
default 1
2014-07-25
6
5
4
3
2
1
213
212
211
210
29
28
1
1
1
1
1
1
1
25
Version 1.0
SFH 7779
ALS_VIS_TL_LSBs register (0x51h)
LSB for the ALS VIS threshold „LOW“.
R/W-Register 0x51
Bit
7
27
default 0
6
5
4
3
2
1
25
24
23
22
21
ALS_VIS_TL_MSBs register (0x52h)
MSB for the ALS VIS threshold „LOW“.
R/W-Register 0x52
Bit
7
215
default 0
6
5
4
3
2
1
213
212
211
210
29
28
INT modes
The Interrupt function compares ALS and PS measurement values with the current interrupt threshold
level. PS and ALS VIS Interrupt status is readable via register 0x4Ah or at the INT pin of the sensor.
The Interrupt persistence function is only valid for PS measurements and is defined at register (0x43h).
The INT pin of the SFH 7779 is an open drain output and should be pulled-up to VI/O by an external
resistor. When VDD is supplied to the sensor the INT pin is per default in high ohmic (inactive) state.
The INT status becomes inactive by writing an INT reset command, reading the INT status register or
performing a software reset.
The INT status stays in its last state when the sensor is set to the standby mode. In the INT active state
„low“ the sensor consumes ~25µA extrra current. Therefore OSRAM recommends to set the INT state
to high impedance before setting the sensor in standby mode.
Below the ALS and PS INT modes are described for the unlatched mode. In latched mode the switching
back to the „inactive“ INT state is depending on an interrupt reset or on reading out the INT status register.
ALS INT mode:
The ALS VIS threshold levels high (register 0x4Fh & 0x50h) and low (register 0x4Fh & 0x50h)
are only valid for the ALS VIS measurement values. The ALS VIS INT mode is fixed and can not
be adapted via register. The thresholds define a window with the following functionality:
ALS INT is active, if the ALS VIS measurement values are outside the window
ALS INT is inactive, if the ALS VIS measurement results are inside the window.
PS INT Modes: Bit of interrupt function setting register (0x4Ah)
00 PS_TH is active only:
The INT state is active, if the PS measurement result is equal or higher than the set PS_TH high
2014-07-25
26
Version 1.0
SFH 7779
threshold.
The INT state is inactive, if the PS measurement result is lower than the set PS_TH high threshold.
PS INT active
PS INT inactive
PS signal
PS_TH
01 PS_TH & PS_TL (PS high & low threshold) are active as hysteresis:
PS_TH and PS_TL are working as a hysteresis. If the PS measurement signal is higher than the
PS high threshold (PS_TH) the INT state is switched to active. If the PS measurement signal is
lower than the PS low threshold (PS_TL) the INT state is inactive. If once the interrupt signal
becomes active, the INT status is kept active until the measurement result becomes less than
the PS_TL register value.
PS INT active
PS INT inactive
PS signal
PS_TH
PS_TL
10 PS_TH & PS_TL (PS high & low threshold) are active as outside detection:
In case of “PS outside detection“ mode interrupt signal inactive means that the measurement
result is within the registered threshold level and the interrupt signal active means that the meas-
2014-07-25
27
Version 1.0
SFH 7779
urement result is out of registered threshold level.
PS INT active
PS INT inactive
PS signal
PS_TL
2014-07-25
PS_TH
28
Version 1.0
SFH 7779
Package Outline
Dimensions in mm / Maße in mm.
The emitter is located in the cavity between pad 4 and 5. The orientation of the sensor can be detected
by the asymmetrical channel on the sensor top side.
For optical aperture design please see page 8.
Package
2014-07-25
Chip on board
29
Version 1.0
SFH 7779
Method of Taping
Dimensions in mm (inch). / Maße in mm (inch).
2014-07-25
30
Version 1.0
SFH 7779
Recommended solder pad design
Cleaning / Washing
In general, OSRAM Opto Semiconductors does not recommend a wet cleaning process for the
components SFH7779 as the package is not hermetically sealed. Due to the open design, all kind of
cleaning liquids can infiltrate the package and cause a degradation or a complete failure of the LED or
ASIC. It is also recommended to prevent penetration of organic substances from the environment
which could interact with the hot surfaces of the operating chips. Ultrasonic cleaning is generally not
recommended for all types of LEDs (see also the application note "Cleaning of LEDs"). As is standard
for the electronic industry, OSRAM Opto Semiconductors recommends using low-residue or no-clean
solder paste, so that PCB cleaning after soldering is no longer required. In any case, all materials and
methods should be tested beforehand in order to determine whether the component will be damaged
in the process.
2014-07-25
31
Version 1.0
SFH 7779
Reflow Soldering Profile
Preconditioning: JEDEC Level 3 acc. to JEDEC S-STD-020-D
OHA04525
300
˚C
T 250
Tp 245 ˚C
240 ˚C
tP
217 ˚C
200
tL
150
tS
100
50
25 ˚C
100
150
200
s 300
250
t
OHA04612
Profile Feature
Profil-Charakteristik
Symbol
Symbol
Pb-Free (SnAgCu) Assembly
Minimum
Ramp-up rate to preheat*)
25 °C to 150 °C
Time tS
TSmin to TSmax
tS
60
Ramp-up rate to peak*)
TSmax to TP
Recommendation
Maximum
2
3
100
120
2
3
Unit
Einheit
K/s
s
K/s
Liquidus temperature
TL
217
Time above liquidus temperature
tL
80
100
s
Peak temperature
TP
245
260
°C
Time within 5 °C of the specified peak
temperature TP - 5 K
tP
s
3
6
K/s
10
Ramp-down rate*
TP to 100 °C
480
Time
25 °C to TP
All temperatures refer to the center of the package, measured on the top of the component
* slope calculation DT/Dt: Dt max. 5 s; fulfillment for the whole T-range
2014-07-25
°C
32
s
Version 1.0
SFH 7779
Disclaimer
Disclaimer
Attention please!
The information describes the type of component and
shall not be considered as assured characteristics.
Terms of delivery and rights to change design reserved.
Due to technical requirements components may contain
dangerous substances.
For information on the types in question please contact
our Sales Organization.?If printed or downloaded,
please find the latest version in the Internet.
Packing
Please use the recycling operators known to you. We
can also help you – get in touch with your nearest sales
office. ?By agreement we will take packing material
back, if it is sorted. You must bear the costs of transport.
For packing material that is returned to us unsorted or
which we are not obliged to accept, we shall have to
invoice you for any costs incurred.
Components used in life-support devices or
systems must be expressly authorized for such
purpose!
Critical components* may only be used in life-support
devices** or systems with the express written approval
of OSRAM OS.
Bitte beachten!
Lieferbedingungen und Änderungen im Design
vorbehalten. Aufgrund technischer Anforderungen
können die Bauteile Gefahrstoffe enthalten. Für weitere
Informationen zu gewünschten Bauteilen, wenden Sie
sich bitte an unseren Vertrieb. Falls Sie dieses
Datenblatt ausgedruckt oder heruntergeladen haben,
finden Sie die aktuellste Version im Internet.
Verpackung
Benutzen Sie bitte die Ihnen bekannten Recyclingwege.
Wenn diese nicht bekannt sein sollten, wenden Sie sich
bitte an das nächstgelegene Vertriebsbüro. Wir nehmen
das Verpackungsmaterial zurück, falls dies vereinbart
wurde und das Material sortiert ist. Sie tragen die
Transportkosten. Für Verpackungsmaterial, das
unsortiert an uns zurückgeschickt wird oder das wir nicht
annehmen müssen, stellen wir Ihnen die anfallenden
Kosten in Rechnung.
Bauteile, die in lebenserhaltenden Apparaten und
Systemen eingesetzt werden, müssen für diese
Zwecke ausdrücklich zugelassen sein!
Kritische Bauteile* dürfen in lebenserhaltenden
Apparaten und Systemen** nur dann eingesetzt
werden, wenn ein schriftliches Einverständnis von
OSRAM OS vorliegt.
*) A critical component is a component used in a
life-support device or system whose failure can
reasonably be expected to cause the failure of that
life-support device or system, or to affect its safety or the
effectiveness of that device or system.
**) Life support devices or systems are intended (a) to be
implanted in the human body, or (b) to support and/or
maintain and sustain human life. If they fail, it is
reasonable to assume that the health and the life of the
user may be endangered.
2014-07-25
*) Ein kritisches Bauteil ist ein Bauteil, das in
lebenserhaltenden
Apparaten
oder
Systemen
eingesetzt wird und dessen Defekt voraussichtlich zu
einer Fehlfunktion dieses lebenserhaltenden Apparates
oder Systems führen wird oder die Sicherheit oder
Effektivität
dieses
Apparates
oder
Systems
beeinträchtigt.
**) Lebenserhaltende Apparate oder Systeme sind für
(a) die Implantierung in den menschlichen Körper oder
(b) für die Lebenserhaltung bestimmt. Falls Sie
versagen, kann davon ausgegangen werden, dass die
Gesundheit und das Leben des Patienten in Gefahr ist.
33
Version 1.0
SFH 7779
Published by OSRAM Opto Semiconductors GmbH
Leibnizstraße 4, D-93055 Regensburg
www.osram-os.com © All Rights Reserved.